First, please let me apologize in advance for what will be a long winded explanation! This is truly a complicated issue, and is not well understood even by many professionals.
The minimum UF rate of 300 ml/hr is something that came about when the dialysis industry converted from “conventional” dialyzers to" high flux" dialyzer in the late 80’ to early 90’s. Because high flux dialyzers are much more porous, there was a concern that dialysate contaminated with bacteria or endotoxin could be pushed into the blood if the dialysate pressure was higher than the blood side pressure.
If you can open the attachment I added to the post, the next part will be easier to understand. Also please note that these numbers are not exact, and will be different for every patient and dialyzer, they are just and example to help explain the concerns.
As blood goes through a dialyzer, the pressure decreases. This means that if your venous pressure is 150 mmHg, then the pressure at the arterial end of the dialyzer is approximately 250 mmHg, and the average blood compartment pressure is 200 mmHg. As dialysate goes through the dialyzer, its pressure decreases as well, but not as much as blood because it is a thinner fluid. If the UF on the machine is off, or set to 0, than the TMP is zero as well. This means that the average dialysate compartment pressure would have to be 200 mmHg to equal the blood compartment pressure. So to maintain an average pressure of 200, the dialysate would enter the dialyzer at 225 mmHg, and exit at 175 mmHg.
Blood and dialysate run in opposite directions in the dialyzer. In this case on the venous end of the dialyzer the pressure on the blood side is 150 mmHg, and the dialysate is 225 mmHg. Fluid will move from a higher pressure to a lower pressure, so dialysate is moving into the blood stream (this is called backfiltration) at the venous end of the dialyzer. On the arterial end, the dialysate pressure is 175 and the blood pressure is 250, so fluid is moving from the blood compartment into the dialysate. With the UF rate (and TMP) at zero, the amount of fluid entering the blood will be equal to the amount being removed, so the is no change inpatient weight.
When many dialysis professionals realized that backfiltration was occurring they were very concerned. The dialysis manufacturers came up with a recommendation of setting a minimum UF of 300 ml/hr to prevent it from happening. His made sense to most of us, and we adopted this as policy and went on with performing high flux dialysis.
In reality, a minimum UF of 300 makes very little difference. In the above example, if the patient was on an F80 dialyzer with a KUF of 60 ml/mmHg/hr the TMP would be 5 mmHg. There would then have to be an average dialysate pressure of 195 mmHg. On the venous end, the pressure would be 220, but the blood side would still be 150, and fluid would still move from the dialysate to the blood. To prevent backfiltration, you would have to have enough TMP to reduce the dialysate pressure on the venous end to the same as the blood side pressure, in this case it would be 150 mmHg. That would make the average dialysate pressure 125 mmHg. With the average blood side pressure at 200 the TMP is 75. That would mean that the UF rate would have to be at least 4500 ml/hr to prevent backfiltration. The reality is that backfiltration occurs during every high flux dialysis treatment. There have never been any problems associated with it, and we have done millions and millions of treatments.
TMP monitors on current dialysis machines are there to monitor gross changes in pressure due to a malfunction of the UF system. They are not precise as far as the display goes: often you have an actual TMP of 5-10 mmHg, and most TMP monitors move in increments of 20 mmHg. They do serve the purpose for which they were intended, but it can be frustrating when you get a lot of TMP alarms. (that is another longwinded explanation!)
I hope this helps!
Curt